SIGforum.com    Main Page  Hop To Forum Categories  The Lounge    Experts Demand Validation of Automatic Blood Pressure Devices
Page 1 2 
Go
New
Find
Notify
Tools
Reply
  
Experts Demand Validation of Automatic Blood Pressure Devices Login/Join 
Member
posted
In line with recommendations by the World Health Organization, the World Hypertension League published a consensus document that calls for the "urgent" strengthening of regulations to ensure that all automated blood pressure measurement devices that are available for sale or are used in health centers are clinically validated. The document was signed by 20 experts from 10 countries. They state that validation ensures device accuracy and reduces the risk of errors and deficiencies in the diagnosis and medical management of arterial hypertension.

"Accelerating the adoption of well-validated automated blood pressure devices in routine clinical use is important for national and global strategies to improve the monitoring of hypertension, cardiovascular disease, and noncommunicable diseases," the authors note in an article published in Journal of Human Hypertension.

A Prevalent Problem
It is estimated that 75% to 80% of the more than 3500 models of automatic devices (sphygmomanometers) for measuring blood pressure that are marketed in the world have not been clinically validated. Validation entails testing in a rigorous and standardized process involving healthy people and patients with hypertension to ensure that the measurements are accurate.

The situation could be even worse in Latin America. "If globally only 20% of the devices are validated, imagine in countries with weaker or nonexistent regulatory systems," signatory Pedro Orduñez, MD, PhD, regional advisor for the prevention and control of cardiovascular diseases of the Pan American Health Organization (PAHO) and leader of the HEARTS initiative in the Americas, told Medscape Spanish Edition.

In Brazil, for example, only 16.7% of 137 devices promoted on popular internet pages had been clinically validated, according to a recent study conducted by Estefanie Siqueira Vigato, MS, health sciences teacher, assistant professor, and doctoral student in nursing at the State University of Campinas (UNICAMP), in Campinas, Brazil.

For Orduñez, the proportion of validated devices must be urgently increased. "It is crazy to continue measuring pressure with equipment that is known to be inaccurate, with the risk for patients and the costs for the health system that this implies," he said.

The consensus signatories cited a Canadian study that demonstrated that nonvalidated devices were associated with clinically significant discrepancies in blood pressure measurement, compared with accurate measurements. The discrepancies have been >5 mm Hg in 69% of patients and >10 mm Hg in 36%.

"At population level, even small errors — for example, 5 mm Hg — in the measurement of systolic or diastolic blood pressure can lead to the misclassification of millions of people," they note.

They add that inaccurate blood pressure measurement can result in unnecessary treatments for patients without hypertension, as well as lack of treatment for patients incorrectly classified as normotensive.

The Proper Procedure
The article's lead author, James Sharman, PhD, deputy director of the Menzies Institute for Medical Research at the University of Tasmania in Hobart, Australia, acknowledged that having a properly validated blood pressure device is just one of many factors necessary for an accurate measurement of blood pressure.

If the professionals performing the blood pressure measurement do not follow the proper procedure, this can lead to misdiagnosis and poor medical management. It is therefore a high priority that healthcare professionals be trained in correct measurement and employ out-of-office assessment when available, such as multiday home blood pressure monitoring or 24-hour ambulatory monitoring," Sharman told Medscape Spanish Edition.

Hypertension is the main risk factor for dying, although the general population and doctors themselves ignore or minimize it, said Orduñez. Many healthcare professionals "overestimate their knowledge and skills in taking blood pressure, even when they know they don't spend time on this matter," he added.

Another factor that can conspire against a greater use of validated automated devices is economic. On average, the cost of validated devices in the international internet market is double that of nonvalidated equipment.

"This is a problem in itself, as it is likely to entice consumers to buy cheaper, nonvalidated devices, especially since the topic of validation is not something people generally know about. So, we need to prevent this from being a barrier for healthcare providers in countries with fewer resources. This problem is being addressed, at least in part, through the PAHO Strategic Fund, which is a mechanism that allows for the large-scale purchase of essential health supplies, including validated blood pressure devices," said Sharman. In 2020, he led the Lancet Commission on Hypertension group, which made a similar claim.

In the new consensus document, the signatories emphasize the need for governments to build a strong regulatory capacity to ensure a quick and smooth transition to the exclusive use of validated automated blood pressure measurement devices. "It is urgent and a technical imperative," they stress.

"Professionals must be educated, but you also have to regulate. It's like traffic: drivers and pedestrians must be educated, but stricter regulations and signs must be put in place for guiding," said Orduñez.

Lists on a webpage on the official PAHO site indicate which models of automated blood pressure measurement devices are clinically validated.

The World Hypertension League document is supported by the following groups and organizations: the Lancet Commission on Hypertension Group, the International Society of Hypertension, Hypertension Canada, the British and Irish Hypertension Society, t Stroke Foundation–Australia, the High Blood Pressure Research Council of Australia, and Resolve to Save Lives.

Orduñez is a PAHO official, but the position of the authors of the consensus document does not necessarily reflect that of the organization. Sharman has disclosed no relevant financial relationships.

Follow Matías A. Loewy of Medscape Spanish Edition on Twitter @MLoewy.
 
Posts: 17703 | Location: Stuck at home | Registered: January 02, 2015Reply With QuoteReport This Post
I Deal In Lead
Picture of Flash-LB
posted Hide Post
That oughta add a healthy amount of money to the price of a blood pressure device.
 
Posts: 10626 | Location: Gilbert Arizona | Registered: March 21, 2013Reply With QuoteReport This Post
Member
posted Hide Post
Whatever it costs, companies who make the devices should foot the bill. Too many so called medical devices are marketed to the public without validation.

The one I use has been validated and was not at all prohibitively expensive.
 
Posts: 17703 | Location: Stuck at home | Registered: January 02, 2015Reply With QuoteReport This Post
Get my pies
outta the oven!

Picture of PASig
posted Hide Post
If the WHO is pushing for this, then I’m against it.

It’s as easy as that.


 
Posts: 35166 | Location: Pennsylvania | Registered: November 12, 2007Reply With QuoteReport This Post
Fighting the good fight
Picture of RogueJSK
posted Hide Post
My doctor's entire practice (a multi-office regional medical group) went back to manual blood pressure readings in the mid-2010s, after using automatic cuffs for a while. At the time of the change, the nurse said doing it manually was much more accurate.
 
Posts: 33463 | Location: Northwest Arkansas | Registered: January 06, 2008Reply With QuoteReport This Post
Member
Picture of konata88
posted Hide Post
Stop buying shit made in prc and that should help a lot.




"Wrong does not cease to be wrong because the majority share in it." L.Tolstoy
"A government is just a body of people, usually, notably, ungoverned." Shepherd Book
 
Posts: 13223 | Location: In the gilded cage | Registered: December 09, 2007Reply With QuoteReport This Post
Knows too little
about too much
Picture of rduckwor
posted Hide Post
quote:
Originally posted by RogueJSK:
My doctor's entire practice (a multi-office regional medical group) went back to manual blood pressure readings in the mid-2010s, after using automatic cuffs for a while. At the time of the change, the nurse said doing it manually was much more accurate.


The problem is that no practice wants to pay people properly trained to take blood pressure in the tried and proven method of auscultation. Much easier to hire a nobody and teach them to push a button. And I agree with Konato88. Stop buying the PRC crap.

RMD




TL Davis: “The Second Amendment is special, not because it protects guns, but because its violation signals a government with the intention to oppress its people…”
Remember: After the first one, the rest are free.
 
Posts: 20426 | Location: L.A. - Lower Alabama | Registered: April 06, 2008Reply With QuoteReport This Post
Ammoholic
posted Hide Post
quote:
Originally posted by ZSMICHAEL:
Whatever it costs, companies who make the devices should foot the bill. Too many so called medical devices are marketed to the public without validation.

The one I use has been validated and was not at all prohibitively expensive.
Companies never foot the bill. It is always rolled into the cost of the product.
 
Posts: 7221 | Location: Lost, but making time. | Registered: February 23, 2011Reply With QuoteReport This Post
Member
posted Hide Post
^^^^^^^^^^^^^^^^^
Of course it is passed onto the consumer. Paying more for a validated product is worth it.
 
Posts: 17703 | Location: Stuck at home | Registered: January 02, 2015Reply With QuoteReport This Post
I Deal In Lead
Picture of Flash-LB
posted Hide Post
quote:
Originally posted by ZSMICHAEL:
Whatever it costs, companies who make the devices should foot the bill. Too many so called medical devices are marketed to the public without validation.

The one I use has been validated and was not at all prohibitively expensive.


Perhaps yes, perhaps no. I spent 10 years working in Medical Electronics, mostly in R&D and any device that had any medical impact on patients was tested by outside agencies to an unbelievable measure.

https://www.fda.gov/medical-de...l%20from%20the%20FDA.

In the U.S., FDA regulates the sale of medical device products. Before a medical device can be legally sold in the U.S., the person or company that wants to sell the device must seek approval from the FDA. To gain approval, they must present evidence that the device is reasonably safe and effective for a particular use.
 
Posts: 10626 | Location: Gilbert Arizona | Registered: March 21, 2013Reply With QuoteReport This Post
goodheart
Picture of sjtill
posted Hide Post
We have an Omron Platinum digital BP cuff. It's readings so far are very close to manual readings with a certified sphygmomanometer. I was trained to do BP readings in a research setting; in the doctor's office hardly anyone makes patient sit down and wait 5 minutes before checking blood pressure. There may be more intrinsic error in that than in the automated devices.

However, years ago I had one of the big "stick your arm in it" devices they had in drug stores compared with our certified devices and the big machines were wildly imprecise and inaccurate.

IMO if you buy a brand-name device that goes on the arm, not the wrist, it will probably be pretty accurate. What's important from the diagnostic point of view is repeated home measurements at rest; ideally bringing the cuff in to the doctor's office to compare with their readings at same time and same conditions.


_________________________
“Remember, remember the fifth of November!"
 
Posts: 18626 | Location: One hop from Paradise | Registered: July 27, 2004Reply With QuoteReport This Post
Go Vols!
Picture of Oz_Shadow
posted Hide Post
Something about the automatics returns insanely high results on me. I have a large Welch Allyn/Tyco manual one with the correct cuff for home.
 
Posts: 17944 | Location: SE Michigan | Registered: February 10, 2007Reply With QuoteReport This Post
Member
Picture of konata88
posted Hide Post
Yes, I've had good experiences with both Omron and Panasonic arm digital cuffs. And both types: 1) measure while pressurizing (faster); 2) measure while depressurizing (slower).




"Wrong does not cease to be wrong because the majority share in it." L.Tolstoy
"A government is just a body of people, usually, notably, ungoverned." Shepherd Book
 
Posts: 13223 | Location: In the gilded cage | Registered: December 09, 2007Reply With QuoteReport This Post
Bookers Bourbon
and a good cigar
Picture of Johnny 3eagles
posted Hide Post
We have a HeartSmart Automatic BP Monitor Model # 01-553. We have taken it to the Primary Care Doctors office and verified it against their manual BP Monitor. Right on the money.





If you're goin' through hell, keep on going.
Don't slow down. If you're scared don't show it.
You might get out before the devil even knows you're there.


NRA ENDOWMENT LIFE MEMBER
 
Posts: 7375 | Location: Arkansas  | Registered: November 06, 2010Reply With QuoteReport This Post
Nullus Anxietas
Picture of ensigmatic
posted Hide Post
quote:
Originally posted by konata88:
Yes, I've had good experiences with both Omron and Panasonic arm digital cuffs. And both types: 1) measure while pressurizing (faster); 2) measure while depressurizing (slower).
No, they don't. They both measure while deflating. See: How do blood pressure monitors really work?



"America is at that awkward stage. It's too late to work within the system,,,, but too early to shoot the bastards." -- Claire Wolfe
"If we let things terrify us, life will not be worth living." -- Seneca the Younger, Roman Stoic philosopher
 
Posts: 26032 | Location: S.E. Michigan | Registered: January 06, 2008Reply With QuoteReport This Post
Washing machine whisperer
Picture of Appliance Brad
posted Hide Post
quote:
Originally posted by rduckwor:
[QUOTE]Originally posted by RogueJSK:
The problem is that no practice wants to pay people properly trained to take blood pressure in the tried and proven method of auscultation. Much easier to hire a nobody and teach them to push a button.
RMD


You're obviously unfamiliar with what they pay EMT-Bs. Wink


__________________________
Writing the next chapter that I've been looking forward to.
 
Posts: 11331 | Location: Willow Fen Farm | Registered: September 17, 2004Reply With QuoteReport This Post
Member
Picture of konata88
posted Hide Post
Well it seems a number of people believe the Panasonic EW3109W measures while inflating. But in any case, seems to work well.




"Wrong does not cease to be wrong because the majority share in it." L.Tolstoy
"A government is just a body of people, usually, notably, ungoverned." Shepherd Book
 
Posts: 13223 | Location: In the gilded cage | Registered: December 09, 2007Reply With QuoteReport This Post
Stupid
Allergy
Picture of dry-fly
posted Hide Post
[/QUOTE]
You're obviously unfamiliar with what they pay EMT-Bs. Wink[/QUOTE]

Or Paramedics either!


"Attack life, it's going to kill you anyway." Steve McQueen...
 
Posts: 7119 | Location: TEXAS | Registered: July 18, 2005Reply With QuoteReport This Post
Muzzle flash
aficionado
Picture of flashguy
posted Hide Post
I use an Omron wrist unit daily to monitor BP. My doctor's office uses manual sphygmomanometers.

flashguy




Texan by choice, not accident of birth
 
Posts: 27911 | Location: Dallas, TX | Registered: May 08, 2006Reply With QuoteReport This Post
I Deal In Lead
Picture of Flash-LB
posted Hide Post
quote:
Originally posted by flashguy:
I use an Omron wrist unit daily to monitor BP. My doctor's office uses manual sphygmomanometers.

flashguy


So does every Doctor I've ever had.
 
Posts: 10626 | Location: Gilbert Arizona | Registered: March 21, 2013Reply With QuoteReport This Post
  Powered by Social Strata Page 1 2  
 

SIGforum.com    Main Page  Hop To Forum Categories  The Lounge    Experts Demand Validation of Automatic Blood Pressure Devices

© SIGforum 2024